[Acoustics] OSW Memo 2006.04 - Availability of the Loop Method
report
Kevin Oberg
kaoberg at usgs.gov
Fri Sep 1 13:42:47 CDT 2006
The memo announcing the availability of the Loop
Method report by David Mueller and Chad Wagner is
below. The pdf version can also be downloaded
from http://hydroacoustics.usgs.gov/policy/
Please note that the step-by-step procedure for
applying the loop method (in the report appendix)
has been slightly revised and is available at
http://hydroacoustics.usgs.gov/policy/. It is
recommended that this revised appendix be printed
out and placed in the subject report. Also, a
copy of the LC program for applying the loop
method can be downloaded from:
http://hydroacoustics.usgs.gov/equipment/software/
Kevin Oberg
============================
August 31, 2006
In Reply Refer To:
Mail Stop 415
OFFICE OF SURFACE WATER TECHNICAL MEMORANDUM 2006.04
SUBJECT: Availability of the report Application
of the Loop Method for Correcting Acoustic
Doppler Current Profiler Discharge Measurements
Biased by Sediment Transport by David S. Mueller
and Chad R. Wagner (Scientific Investigations
Report 2006-5079) and guidance on the application of the Loop Method
This memorandum (1) announces the availability of
the report Application of the Loop Method for
Correcting Acoustic Doppler Current Profiler
Discharge Measurements Biased by Sediment
Transport (SIR 2006-5079), and (2) provides
guidance on the application of the loop method.
The loop method may be used to correct a
discharge measurement made with an acoustic
Doppler current profiler (ADCP) that is biased
low by sediment transport, often referred to as a
moving bed, and to make the required moving-bed
test (OSW Technical Memorandum 2002.2). When a
moving bed is present, the use of a
differentially corrected global positioning
system (DGPS) for the navigation reference is the
preferred method of making a discharge
measurement with an ADCP. However, there are
situations where a DGPS is unavailable or cannot
be used reliably. In these situations, the loop
method is a valuable alternative for correcting
biases in discharge measurements caused by a moving bed.
The loop method must be applied properly or it
may produce incorrect results. Anyone planning
to use the loop method should read and follow the
subject report (SIR 2006-5079), which describes
the procedures, limitations, and uncertainties
associated with the loop method. Some key aspects
of the loop method are summarized here.
1. The ADCP compass must be properly calibrated
using the manufacturers internal compass
calibration routines. For Rio Grande
instruments, the compass should be evaluated
after the calibration is completed. It is
recommended that the compass error during
evaluation should be less than 1-degree.
2. The navigation reference must be set to
bottom track in the ADCP data-collection software (WinRiver or RiverSurveyor).
3. Bottom track must be maintained throughout
the loop. Loss of bottom track during the loop
will cause inaccuracies in the computed
moving-bed velocity. The amount of bottom track
data that can be lost without significant impact
on the method is difficult to specify. The
hydrographer must consider how much data are lost
and if the lost data are accurately represented
by adjacent data. If it is determined that lost
bottom track is adversely impacting the loop
method, the loop method may not be appropriate
for that location and flow condition and another
method of determining the moving-bed bias or
means of measuring the discharge may be needed.
4. The loop should begin with the boat located
at a fixed, nonferrous marker. The boat is
maneuvered to make the discharge measurement at a
uniform speed back and forth across the channel,
including while turning the boat at the far
shore. The boat must return to the exact same
starting point marker. Uniform speed is required
to obtain a spatially uniform sampling of the
moving-bed conditions throughout the cross
section. If boat speed varies, the moving-bed
computation will be biased by the part of the
cross section with the most data.
5. The duration of the loop should be 3 minutes
or greater, and the boat speed should not exceed
1.5 times the mean downstream water velocity.
When appropriate, using the loop method as a
moving-bed test has the advantage of measuring
bed conditions throughout the part of the cross
section that can be directly measured with the
ADCP. The loop method, therefore, results in a
more representative moving-bed test than a
single-location stationary moving-bed
test. Either a loop test or stationary test can
be used to satisfy the moving-bed test
requirement described in OSW Technical Memorandum 2002.2.
The subject report describes an analysis of the
uncertainty associated with the loop method and
concludes that the uncertainty is 0.02 ft/s. A
measured moving-bed velocity of less than 0.04
ft/s (2-sigma) may be caused by method
uncertainty rather than sediment transport. When
using the loop method, a measured mean moving-bed
velocity of at least 0.04 ft/s indicates the
presence of a moving-bed. If the measured moving
bed velocity exceeds 0.04 ft/s, the ratio of the
mean moving-bed velocity and mean water velocity
should be computed by dividing the mean
moving-bed velocity by the mean water
velocity. If this ratio is greater than 0.01,
the apparent bed movement will cause at least a
1-percent negative bias in the computed discharge
and a method that accounts for or corrects for a
moving bed should be used. NOTE: The guidance
provided in this paragraph supersedes the
guidance described in Step 4, Processing for
Moving-Bed Test, in the AppendixStep-by-Step
Procedures for Using the Loop Method of SIR 2006-5079.
EXAMPLE 1: The mean water velocity is 2 ft/s and
the loop method measured a moving bed of 0.03
ft/s. It can be assumed that there is no moving
bed condition because the moving bed velocity is less than 0.04 ft/s.
EXAMPLE 2: The mean water velocity is 2 ft/s and
the loop method measured a moving bed of 0.05
ft/s. The ratio of the mean moving-bed velocity
and mean water velocity should be computed by
dividing 0.05 ft/s by 2 ft/s to yield 0.025,
which is greater than 0.01 and indicates that a
method that accounts for or corrects for the moving bed should be used.
EXAMPLE 3: The mean water velocity is 6 ft/s and
the loop method measured a moving bed of 0.05
ft/s. The ratio of the mean moving-bed velocity
and mean water velocity should be computed by
dividing 0.05 ft/s by 6 ft/s to yield 0.0083,
which is less than 0.01 and thus no significant
bias in discharge will be caused by a moving bed.
The mean moving-bed velocity measured by the loop
method can also be used to correct the measured
discharge for the bias caused by the moving
bed. The loop method correction only adjusts the
discharge; the measured velocities are still
biased low. To use the loop method to correct
the measured discharge, the discharge should be
measured in accordance with OSW Technical
Memorandum 2002.02 and the moving-bed velocity
measured using the loop method as described in
the subject report (SIR 2006-5079). The subject
report presents two methods, mean and
distributed, for applying this correction to the
discharge measurement. The mean-correction
method, which can be easily computed by hand,
uses the mean moving-bed velocity and the
cross-sectional area perpendicular to the flow to
compute the discharge under calculated due to the
moving bed. This discharge is simply added to
the measured discharge to obtain the final
corrected discharge. The second correction
method distributes the mean moving-bed velocity
across the cross section based on the near-bed
water velocity in each ensemble or profile. This
distributed correction should provide a more
accurate correction for those cross sections that
have significant spatial variations in depth and
moving-bed velocity. Computations needed to apply
the distributed correction method are best
accomplished by a computer program such as the program Loop Correction (LC).
Loop Correction is a computer program written to
determine if a moving bed is present from the
loop method data and to correct the measured
discharge using the distributed correction
method. The computer program reads ASCII files
in the standard ASCII output format from Teledyne
RD Instruments WinRiver. SonTek/YSI
RiverSurveyor can also output files consistent
with this format. The LC program first requests
the user load the loop data file. The mean
moving-bed velocity is computed and the direction
checked to verify that it is in the upstream
direction. The criteria for determining if the
data reflect a moving bed are applied and
reported to the user. If a moving-bed is
detected, the user can select the files
(transects) for the discharge measurement. The
program will read the files and compute the
corrected discharge. The results of the entire
process are displayed for the user and can be
saved to a file and printed for inclusion in the measurement archive.
A copy of subject report, SIR 2006-5079, can be
downloaded from: http://pubs.usgs.gov/sir/2006/5079/
A revised Appendix for the subject report (SIR
2006-5079) of the step-by-step procedures for
using the loop method that are consistent with
this memorandum can be downloaded
from: http://hydroacoustics.usgs.gov/policy/.
It is recommended that this revised Appendix be
printed out and placed in the subject report.
A copy of the LC program can be downloaded from:
http://hydroacoustics.usgs.gov/equipment/software/
If you have any questions or comments about the
policies and guidance in this memo, please
contact David S. Mueller (dmueller at usgs.gov),
Kevin Oberg (kaoberg at usgs.gov), or the OSW
Hydroacoustics Work Group (hawg at simon.er.usgs.gov)