When you say you are looking at Beam 1 and Beam 2 velocities, do you mean
you are looking at the x and y components of velocity (x being red and y
being blue)? From your description, it does seem possible that the
Argonaut has been rotated as Liz suggested.
Is the instrument located near or on a river bend...are the water levels
much different this year than last year?
Are there nearby tributary inflows upstream of the SonTek?
Do the horizontally averaged velocities also differ from the previous
installation or only the last two cells?
What frequency is the unit? How long is your sampling range?
You state that only the last two cells of the multi-cell data seem to be
affected...that may be that the Argonaut is rotated in pitch or roll.
Are the noise level of the two instruments similar? Are the return signal
strengths above the instrument noise.
Are the suspended solids concentrations different from one year to the
next?
I seem to have more questions than answers.
If you want to discuss this, give me a call,
Victor Levesque
U.S. Geological Survey
Florida Science Center
St. Petersburg and Tampa, FL
727-803-8747 extension 3085
813-975-8620 extension 167
"Keith Ging" <Keith.Ging at lcra.org>
Sent by: acoustics-bounces at simon.er.usgs.gov
08/30/2006 07:22 PM
To
<acoustics at simon.er.usgs.gov>
cc
Steven Clark <Steven.Clark at lcra.org>, Matthew Ables
<Matthew.Ables at lcra.org>, Gene Taylor <gene.taylor at lcra.org>
Subject
[Acoustics] Question on Sontek SL beam vel's
I wanted to run this across everyone to see if my line of thought was near
correct. I've recently found that our rating has changed at one of our
index-velocity sites based on 4 discharge measurements this year. We
cannot find any visual clues to suggest the channel or vegetation has
changed.
After digging into the data through ViewArgonaut, I noticed two things
that stood out.
1. The multi-cell data shows the velocities of the last two cells are
fluctuating more than usual at very slow vel's (<0.20 ft/sec) as compared
to the previous 2 years (2004-2005). However, the diagnostic data does
not show anything in the path of the beams based on signal strength and
SNR's. Visual observations have found nothing in the channel to be
different.
2. I then looked at the beam velocity data and found the vel's of beam
1 & 2 to roughly equal one another fairly consistently in 2004 (the
magnitude is near the same but one is negative and one is positive). This
changed in 2005 and 2006. Beam2 is about 0.10 ft/sec slower than Beam1 at
the slower vel's. Many times Beam2 is half of Beam1 vel's. Could this
indicate the SL has moved slightly? The SL was replaced during the
off-season between 2004 and 2005. We do have it marked to prevent this
from happening. We also recently shot levels to make sure the SL had not
changed in elevation which it had not. I can only guess it may be turned
slightly when it was replaced and we will be checking this soon.
Please let me know your thoughts. I'm interested in how others have used
the beam vel's to diagnose problems or any other parameter logged by the
SL.
Thanks,
Keith Ging, P.G., Senior Hydrologist
River Operations
Lower Colorado River Authority
3505 Montopolis Dr, Austin, TX 78744
Phone: 512-356-6009
Toll Free: 1-800-776-5272 ext. 6009
Fax: 512-356-6059
Email: kging at lcra.org
_______________________________________________
Acoustics mailing list
Acoustics at simon.er.usgs.gov
http://simon.er.usgs.gov/mailman/listinfo/acoustics
-------------- next part --------------
An HTML attachment was scrubbed...
URL: http://simon.er.usgs.gov/pipermail/acoustics/attachments/20060831/5fef098d/attachment-0001.html
More information about the Acoustics
mailing list